Status Error

Up to Velocimeters - single point

Status Error

Posted by raymond guang at November 16. 2009

HI

i have used vectrino velocitmeter for a few weeks now. but by last week, the velocimeter keeps give me error status

please have a look of the attachment image (in the words file), how can i fix it?

am i using it for too long? because it was very warm in the lab during summer time in australia

 

and also, my probe doesn't sense the temperature as well

how am i be able to fix it?

 

many thanks

 

ray

Attachments

Re: Status Error

Posted by Marius Mathisen at November 17. 2009

Hi Ray,

 

I see from the screenshot that you use a very old firmware. You should upgrade to the latest firmware 1.29. You can download it from http://www.nortek-as.com/en/support/software. You should also download the latest Vectrino PC software. Please do this first and report the results!

 

Best regards

Marius

Re: Status Error

Posted by raymond guang at November 18. 2009

Thank you Marius for your help

and also i want to ask

sometimes i've got negative velocity readings

i tried to measure the velocity for a period of time, thought the negative sign may disappear. however, the negative sign is still there no matter how long i wait. but if i change the height of the probe to another different height, the velocity readings are more logical

what could be the problem?

please see attachment

Ray

Attachments

Re: Status Error

Posted by P.J. Rusello at November 18. 2009

Ray,

In your attached plot your correlation values are extremely low. In general the correlation values should be upwards of 90%. Please check your Nominal Velocity Range setting in the Deployment Planning dialog and ensure it is high enough for the flow you are measuring. In this same dialog on the right hand side, two velocity ranges are reported, one each for horizontal and vertical velocities. The Nominal Velocity Range setting will cause changes in the reported horizontal and vertical ranges. Make sure these are above the expected flow velocities or you will not measure usable, high quality data. I would suggest moving up to a higher Nominal Velocity Range to start.

 

P.J.

Re: Status Error

Posted by raymond guang at November 19. 2009

Thank you very much !

also, because of this velocimetre is a few years old, and my supervisor couldn't locate the sofeware licence key.

am i be able to question the supplier for the licence key?

 

if anyone can help me on this

 

ray

Re: Status Error

Posted by P.J. Rusello at November 19. 2009
Please locate the office the instrument was purchased from and email a request to them. They will need the probe head serial number and Vectrino body serial number to supply you with a license key to upgrade the firmware.

Re: Status Error

Posted by raymond guang at November 20. 2009

Thanks Peter

i have already emailed them the serial number and the probe number

so do you think the error i am having (see the very first attachment) is related to the firmware version?

see if i upgrade the firmware, the error status will not appear?

 

ray

Re: Status Error

Posted by P.J. Rusello at November 20. 2009

I would recommend updating the firmware and seeing if this solves the status error message. Quite a bit has changed in the six firmware releases since the version your instrument currently has and this could resolve the issue.