HR 1MHz minimum cell size

Up to HR-Profilers

HR 1MHz minimum cell size

Posted by dekeete at February 16. 2011

Hi there,

 

I experienced this situation:

when using the burst sampling mode, the minimu cell size must be larger than 25mm, however when using non-burst mode, it can be setup to 15mm. And another thing is that when set a very small cell size, it will give the infromation "Invalid cell size Valid range is 20-300mm". I don't understant why I can set 15mm in non-burst mode?

 

Another thing:

when using the following setup: Meas/Burst: 60 Average Interval: 30, each profile will only have 3 Ensemble to average for each profile,

while when using the following setup: Meas/Burst: 60 Burst sampling, Samples can be reach 58.

 

Would you like to give me some explanation. Looking forward your answer. Thanks.

 

With best regards

 

Yuanye Wang

Shanghai Estuarine and Coastal Reseach Center

Re: HR 1MHz minimum cell size

Posted by Jonas Røstad at February 16. 2011

Dear Yuanye Wang,


I cannot recreate the errors described. Please make sure that you have the newest software available (currently 1.08 available at http://www.nortekusa.com/en/support/software)

If you still experience the problems, please tell me and attach the .dep file for this. The .dep file is created when choosing File->Save in the main menu.


Best regards

Jonas Røstad

 