AquaDopp HR 2MHz minimum cell size
Posted by NGUYEN TAT THANG at Tuesday 20:07
Dear All
I just found that eventhough I set the minimum cell size to 7mm in the AquaDopp HR 2MHz, it still appears to be 10mm in the *.hdr result file.
Could anybody give me advice about that? Thank you so much.
Thang
Posted by Sven Nylund at Wednesday 11:36
Dear Thang,
Have you verified that you use the latest software version (which can be downloaded from our support page)?
If you have the latest software version, can you upload the header file so I can take a look?
Best regards,
Sven Nylund
I am using the AquaProHR High Resolution, version 1.05 Beta5.
Could you please have a look at the attached file to see if I have had any problems with the setting?
Thank you so much.
Thang
Dear Thang,
Your setting is correct, this is caused by a round-off error in the software which is also present in version 1.06. Although the value is output in mm in the header file it is treated as a cm value so all cell sizes from 7-14 mm are output as 10 mm. I will pass the information on to our software programmer so that it is fixed in the next release.
Best regards,
Sven Nylund
Dear Sven Nylund
Thank you so much for your explanation.
Best wishes
Thang